Also is it even necessary to use purge flow?
Advertisement
Discussions about GC and other "gas phase" separation techniques.
I'm using an Agilent 6890. Solvent is Water. Analytes are MeOH, ACN, IPA and N-methylpyrollidone. Injection volume in 0.1µL. Column flow is 4.6 ml/min.You are using the 7890 right? For splitt-less i usually have the purge flow set to 50 mL/min. Probably the most crucial part is the split-less time you chose. This can and will greatly affect your analyte response and may cause irregular peak shape if it is too long. If you are analyzing low BP compounds I think 0.5 min or less is a good start. But I usually optimize it by starting with a really low split-time and move up until you stop seeing either increase in response or poor peak shape.
What is your injection volume and solvent? What's your column flow rate (mL/min)?