Hi to all,
I'm running 8260c on a 6890/5973N + Tekmar Stratum concentrator. i have problem with my purge & trap. i have follow the same instruction that i get from teledyne instrument web. my instrument have inject the vial but doesn't suck the sample (which i use the new septa, to avoid pressure decreased). can you help me to solve this problem, because before this i used tekmar 3100, that i can adjust their pressure - to make sure the autosampler sucks the sample and go to sparger. but stratum ptc, doesn't have criteria to adjust its pressure. It is because there are something that blocked the vial and its lines/tubing? could anyone help me in solving this problem, please?
On my system ( Stratum , Aquatek 70) the sample delivery is controlled from the auto sampler ( Aquatek 70).
You are using Teklink, correct. Make sure that the autosampler is enabled in your Teklink method.
yup, i'm already connect it with the autosampler & using this autosampler, it's moved and inject in to the sample but it is doesn't suck the sample. i'm try&error every possible things through the software but the same problem still appear. should i change it's tubing? i doesn't want to try it because afraid it will damage other parts in the autosampler.
The sample is transfered with gas pressure. Do you have any pressure showing at the autosampler. I think 20 psi is suggested. Check to see if your sample needle is plugged.

Also check the needle insertion depth. It has to go in enough so both holes are in the vial. There is a hole near the top that supplies pressure that must be under the septa inside the vial in order for the sample to be transfered.
Ok. noted. tq.
6 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry