Disabling detector on HP 5890 series II
Posted: Fri Apr 20, 2012 2:18 pm
by shoom
I have an HP 5890 that I suspect the detector board is faulty. On start up it gives the error Fault : DETB TEMP RDG. I have checked the thermocouples and they are all ok. Is there a way to just disable it so I can use the instrument? Will removing the board do this?
Thanks in advance.
Shoom
Re: Disabling detector on HP 5890 series II
Posted: Fri Apr 20, 2012 3:42 pm
by Consumer Products Guy
That's what I would try, then the unit will "think" that it only has one detector in there.
Maybe you could swap your "good" board into that spot to see if that proves to you that the board itself is the issue, and not just a bad FID heater or sensor.
Parts to fix, maybe even used FID board, should be available, maybe through 3rd party vendors or some (no names) who keep used parts around.
Posted: Fri Apr 20, 2012 4:22 pm
by GasMan
It is a long time since I used a 5890, but I seem to remember that the detector boards only handled the detector signal. The detector temperature was controlled directly from the main board.
Gasman
Posted: Tue Apr 24, 2012 12:34 pm
by shoom
Thanks for the replies. The board has been taken out and it still gives the same error. Back to the drawing board. :cry:
Posted: Mon Apr 30, 2012 2:14 pm
by AICMM
shoom,
You need to remove the RTD connector and heater connector for that detector from the main board. They are located at the upper right hand side of the mother board. The error you are getting is not related to the FID electronics, it is related to the RTD.
Best regards,
AICMM