Ok next step -- how do you plan to transfer sample from the sample point to the GC?
If you transfer liquid, depending on the distance and line size, your hold up may be quite large. If you need to use smaller amounts of sample, or if you need quick GC response to changes in the sample, you may want to consider vaporizing your sample and transferring it to the GC in a heated sample line.
A system that I am setting up at this very moment uses a heated sample line to carry condensables in the vapor phase to the GC. We have a 5 meter line and low liquid flow rates (~0.15mL/min) so this adaptation was pretty much forced... I am modifying the GC injection zone (basically adding another oven to the GC ^_^) so I can keep the injection valve, sample loop, and transfer lines warm with no cold spots. We have used a do-it-yourself approach, using thermocouples and heat tape/wire from cole-parmer and using Solo controllers from Automation Direct; there may be a pre-built solution available as well.
Once you get the sample to the GC, you need a pneumatic actuated valve. For a standard set up (no column switching or backflushing) you want a six-port injection valve -- check out
VICIs selection here. The valve will use solenoid to control the nitrogen/air and turn the valve; the solenoid is controlled by a relay usually in the GC but possibly controlled by the computer or other hardware.
The valve can replace one of your injector ports or be used in series with one of them. There are advantages either way.
I don't have any experience with the clarus 500, but I have used turbochrom (totalchrom's ancestor) plus external digitizer boxes to do automation of valves, relays, method start, column switching, etc. Hopefully someone else can jump in and give you more specific advice.
Good luck,
aidan