Our setup:
GC: Agilent 6850
Column: HP-5
Inlet Liner: Single taper with deactivated glass wool
Sample: Acetone, IPA, hydrocarbons
1) Should I worry about IPA "sticking" to the glass wool in the liner? If so, should I choose a new liner or derivatize? Because the sample is to be washed through silica gel prior to GC analysis, I was considering derivatization as a possible step for the chemist running the sample through the silica as a way to prevent the IPA from sticking to that as well.
2) I can't separate acetone and IPA for whatever reason on this column. Without getting a new column, derivatization with BSA or BSTFA is my solution. Neither derivatizing agent will affect an HP-5 column, correct?
3) If the silver bullet to my analysis is derivatization, is there a way to calibrate a GC using a derivatized sample, or is the error going to be too large to give meaningful results (i.e. is there anyway of quantifying the extent of the derivatization process)?
-My idea was to first calibrate the GC for IPA, then derivatize and check the cal curve for amount of IPA calculated and if it reads none, to say that when using BSA as solvent, the IPA is completely derivatized. When I make that conclusion, I can derivatize a few different concentrations of IPA and then use that calibration with my actual samples. This hinges on IPA making it through the liner wool obviously.
Thanks for the help. I always appreciate the comments I get from you guys as you've all helped me many times in the past. ^_^