Since the problem is worse with the most volatile component when the sample is present, it could be a matter of non-homogeneity of MtBE content in the sample which causes variable amounts of MtBE measured in the preparations containing the sample.
Removing all solvents from a dissolved sample first and then spiking portions of this solution with the solvents of interest should show if this is the problem.
If the low RSD results were performed with the same method that the higher RSD results were achieved, then this indicates that the presence of the sample is the cause.
The factors which must be optimized for the sample preparations are time, temperature, and pressure. These seem to be different than for the blank preparations without the sample.
You may have too short of an equilibration time after pressurization of the vial, or not enough sample is generated to properly flush out the empty sample loop.
Good luck,
Rod
I'm using an standard Agilent G1888. Do you have any guildlines on what these parameters should be? The method now is:
aux pressure 14 psi
Vial Eq.:15 mins
Vial Pressuration: 0.15
loop fill: 0.50 min
loop eq.: 0.06min
inject:0.50min
oven temp:90°C
loop temp 180°c
transfer temp 180°C
One more question that come to mind. I'm using a salting out technique for the prep, but the sample is difficult to get into solution and may not always be in solution prior to the salt addition. Might this cause a problem?