1,2,3-Trichloropropane Analysis by 524.2
Posted: Tue Apr 10, 2012 6:21 pm
by 01310040231
I am passing all the analytes of PE samples by epa method 524.2 except for 1,2,3-Trichloropropane which is reading 30% lower than the true value (I tried another trial study and it was reading low as well!). The calibration looks great, CCV is almost perfect, and the second source standard looks great. I appreciate any suggestions.
Re: 1,2,3-Trichloropropane Analysis by 524.2
Posted: Tue Apr 10, 2012 7:48 pm
by nvalentine
What level is your PE concentration relative to the CCV and 2nd source?
Posted: Tue Apr 10, 2012 8:00 pm
by 01310040231
The CCV is 25 ug/l and the PE is 18ug/l. Thank you Nathan.
Posted: Wed Apr 11, 2012 12:26 pm
by JI2002
Does trichloropropane coelute with any other compound(s) in the standard?
Posted: Wed Apr 11, 2012 12:39 pm
by Bigbear
Is this a new issue for you? Has 123TCP ever worked? If yes, what has changed? How old is your trap?
I do 524.2 and do not have issues with 123TCP. My %RSD from my latest calibration is 7.8.
Posted: Thu Apr 12, 2012 2:58 pm
by 01310040231
TCP is not coeluting with any other compound. The RSD is 6.1% and trap is about two months old. We would like to be able to report this compound on our reports but we either fail or barley pass the PEs for this compound. Thank you all.
Posted: Fri Apr 13, 2012 3:44 am
by bisnettrj2
Are you getting PTs from one provider only, or multiple? Have you tried purchasing a third standard to confirm that you haven't somehow purchased two standards that agree but are actually low (shot in the dark on that one)?
Posted: Fri Apr 20, 2012 12:38 pm
by Bigbear
I just saw something in my set up. Make sure you are integrating correctly. In my system 123 TCP comes out just after bromo benzene which has 75.