Headspace - I am new to headspace testing and I am tryng to develop a method for MHE as a means for quantifying sesamol in plastic film. In theory each subsequent injection should yield a smaller peak area, which is necessary to establish a regression and get slope and so forth to plug in to formulas to get amount in vial etc. If r value is poor - that would indicate the sample is not reaching equilibrium (btwn matrix and headspace or gas phase). My problem is I am getting identical peak area for each injection. What does this scenario mean??? I know sesamol is vary volatile - how do I change my method to get MHE to work with sesamol?
Most likely your sample is too big and the headspace is becoming saturated before an equilibrium is reached.

Every time you reach what you think is equilibrium it is actually only a saturated system. Continued extractions continue to reach saturation of the system so your peak size remains constant.

I would try reducing the size of your sample.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry