I'm analyzing environmental samples for Chloroform, 1,1,1-Trichloroethane, and Carbon Tetrachloride. For each sample 3L of air is pulled through a Carbotrap 300 thermal desorption tube. We have a Gerstel thermal desorption autosampler which goes into a Gerstel CIS inlet into the Agilent 7890 GC, with an Agilent uECD detector.
I have an issue with an unknown peak coming out on top of 1,1,1-TCE. It's something from the sampled air, as it does not show up in any blanks or lab spikes. I am looking for help as to what this could possibly be. According to the Restek application on this column (http://www.restek.com/chromatogram/view/GC_EV1169), it could be dibromofluoromethane? My colleagues tell me it's unlikely that we'd see that in environmental samples. I was hoping for any other suggestions as to what it could possibly be, or really any insight as to what I can do. Here is what the relevant portion of the chromatogram looks like:

Background info:
CIS: Tenax TA packed liner
Trap at 5C, then 12C/sec to 275C, hold 3 min
Oven: 40C for 2 min, 5C/min to 100C, 120C/min to 240C, hold 4 min
PTV inlet:
Mode - Solvent Vent
Pressure - 11.905 psi
Gas saver - off
Purge flow to split vent - 20mL/min at 0.01min
Vent flow - 50 mL/min
Column: Restek Rxi-624sil MS
30m x 250um x 1.4 um
uECD:
240C
Makeup: Ar/CH4 40mL/min