- Posts: 22
- Joined: Tue Dec 13, 2011 5:37 pm
I am currently calibrating a YoungLin 6100GC for FAME analysis of C4:0 through to C24:0 including common isomers such as C18:1c9 C18:2c9,12 and C18:3c9,12,15. Nothing out of the ordinary for FAME analysis. I am experiencing a problem with the C16:0 to C20:0 calibration however as the responses are not linear, not even close.
Brief overview of the set up is
Split injector, 1ul Inj, 50:1 split, 250C,
Column DB-23 30M x 0.32mm x 0.25um
Oven Program start 100 slow ramp to 240
FID Detector 250C
Hydrogen Carrier Gas from Generators
I am using fresh, pure standards from a reputable company and preparing 0.25mg/ml, 0.50mg/ml, 1.00mg/ml and 2.00mg/ml standard solutions in Iso-Octane. For C4:0 to C15:0 and C22:0 to C24:0 linear responses are seen though the C22:0 and C24:0 responses are very much reduced. The C16:0 to C20:0 responses exhibit a plateu effect to 2.0mg/ml then see a somewhat linear rise in response to 5.0mg/ml (examples below, first column after name is the RT, then 0.25mg/ml response through to 2.0mg/ml response mV.s)
FAME C12:0 4.005 76.323 144.532 271.643 507.382
FAME C16:0 6.130 82.430 150.352 268.325 348.683
FAME C17:0 6.750 82.370 143.618 180.634 243.405
FAME C18:0 7.450 75.485 101.217 122.042 177.280
FAME C18:1 7.630 81.805 130.967 153.704 207.213
FAME C18:2 8.010 84.652 137.970 166.071 222.016
FAME C18:3 8.510 86.241 133.995 164.447 211.727
FAME C20:0 9.040 40.857 45.593 55.789 84.901
FAME C22:0 10.935 16.441 24.124 36.026 64.605
FAME C24:0 13.780 8.582 13.919 29.163 56.145
Is there an obivous cause for this as i have not encountered it before? I believe the system set up to be fine and doubt it is due to analytical error. Any help would be appreciated. Thanks