Optimum Retention for a Column (based on size)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Dr. Neue made the comment below:

viewtopic.php?t=11144&highlight=retention
PostPosted: Mon Aug 17, 2009 2:55 pm
Dancho is right: the retention is way too low. Add some more water to the mobile phase! You want a retention of at least around 10 minutes with the flow rate and the column that you have.
-note- flow rate = 1ml/min, column Vydac C18 218 TP (4.6 * 250 mm)

My question is, how does someone know what the retention of a compound should be based on column dimensions? Is it a calculation from a specific capacity factor?

For my newest method, I have a capacity factor of 1.3 so far and really I don't know if that is optimum. I can think of how it might be a problem when I enter into the sample prep development phase, but until then I want to make sure I have the basics down.

Thank you all in advance for your infinite HPLC wisdom! :)

Dear Kerri

for isocratic separations one preferably wants to tune your components into a retention window of about 1 < k < 10 (maybe up to 20).

In combination with the dead time t0, which depends on the colum dimension and the flow rate, one can calculate the corresponding retention time window: tR=t0*(k+1)

The dead time t0 can be estimated by dividing the free column volume by the flow rate:
t0=1/6*d^2*Pi*L/F
(d and L in [cm], Flow in [ml/min], t0 in [min]).
Note: the free column volume can be estimated as about 2/3 of the cylindrical volume of the column, this means that only 1/3 is occupied by the skeleton of the stationary phase.

In the given example, the t0 of the method would have been about 3 min, which turns into a retention factor of only 0.5 for a compound eluting at 4.4 min, whereas 10 min corresponds to a k = ca 2.3

Hollow,

Thank you very much!

Kerri
Hi,

I saw an article where the authors report a dead volume RT of 4.9 for the column

ZIC-HILIC 150 mm x 4.6 mm, 5 μm column (Merck Sequant), The flow rate was 300 μL/min

but with the formulaabove I get 5.53, m I missing some thing? Do I have to take the 5um particle in consideration?
You're not missing anything, but that forumula is only an approximation (give or take 15%). So the measured and predicted are in the same ballpark.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry