Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello,

Recently I performed an infusion with high amount of analyte ~10,000ng/mL at higher flow rate 30uL/min.
In my MRM method I have noticed that the background signal of mobile phase for my analyte, the one that was infused, is high ~1500-2000counts, whereby the I.S. which was not infused gives normal counts ~20-30. I have waited about 1 hour and have noticed counts dropped a bit for the analyte to 300-500, but is still much higher than normal (38) counts Any ideas, or advice as to and how to flush/purge the MS system? I think it is just a waiting game to let the N2 gas clear the ion transfer capillary and quadruples. I have also come to realize next time I will infuse only ~2,000ng/mL at a lower flow rate of 10uL/min.

Thanks.
Hello,

Recently I performed an infusion with high amount of analyte ~10,000ng/mL at higher flow rate 30uL/min.
In my MRM method I have noticed that the background signal of mobile phase for my analyte, the one that was infused, is high ~1500-2000counts, whereby the I.S. which was not infused gives normal counts ~20-30. I have waited about 1 hour and have noticed counts dropped a bit for the analyte to 300-500, but is still much higher than normal (38) counts Any ideas, or advice as to and how to flush/purge the MS system? I think it is just a waiting game to let the N2 gas clear the ion transfer capillary and quadruples. I have also come to realize next time I will infuse only ~2,000ng/mL at a lower flow rate of 10uL/min.

Thanks.
Carryover can happen in a lot of places. Try replacing any tubing or fittings that you used for the infusion with new ones. You could also try flushing the system with LC grade isopropanol.
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry