Configuring Agilent 1050 and 1100 modules together...
Posted: Tue Apr 03, 2012 4:48 am
Previously, an HPLC was configured as three 1050 modules:
autosampler, pump & VWD. The system was operating with Chemstation
A.10.02. The VWD died and we replaced it with an 1100 G1314
VWD. Just like the old VWD, the G1314 connects via HPIB. We hooked
everything up to the G1314 the way it was connected to the previous
1050 VWD. We went into the GC Chemstation Configuration Editor, it
was selected for modular 3D LC system, and we started by removing the
1050 VWD detector. We noticed there were no specific modules listed
for the 1100, but rather an 1100 System Access option. So, we
selected that and kept the GPIB at address 0. We then tried to
connect to Chemstation, and it asked which 1100 module was connected.
We selected the G1314, it came up as offline after we selected it, and
after several minutes Chemstation loaded. On this first connect, we
were able to communicate with the autosampler and pump, but it stated
there was a power failure with the G1314. We rebooted the software
and G1314, but still couldn't connect. We then went into Visa
Assistant. We can see a bunch of activity listed for GPIB0:16 (1050
pump), GPIB0:18 (1050 autosampler) and GPIB0:24 (INSTR - we assume
this is the G1314). When we open the IO CONFIG, we noticed the serial
number on the 82350 PCI-GPIB card had changed after connecting it to
the G1314. We have tried naming the SICL Interface both hp82341and
hpib7, the visa interface name we have left at GPIB0, logical unit = 7
and Bus Address = 30.
We simply cannot get Chemstation A.10.02 to talk to the G1314 VWD at
all, no matter what we try. It reads offline. And now, it also says
that the 1050 autosampler & 1050 pump have power failure when
Chemstation finally opens.
Do any of you have any thoughts as to what the solution might be? Is it
possible to control an 1100 detector with two 1050 modules in A.10.02
Chemstation? Any advice would be greatly appreciated! Thanks so much!