MMJ88 » Wed Dec 07, 2011 5:47 pm
The ECD is running at 240C. Is that abnormally low? I apologize for my inexperience with this, I have only been using MS detectors until now.
The sample tubes in the actual study are sampled from a vapor source, but these tubes for the calibration are simply spiked on the head of the tube with 1uL of methanol with the appropriate concentrations of analytes, then room temperature nitrogen is run through for 3 min at 50mL/min.
The liner is a glass bead one.
The CIS is another can of worms entirely. I'm not using it right now. I had (what looked to be) a pretty good CIS method. The peaks were completely separated, sharp, and symmetrical. However, the response would vary wildly. I could run the same spike multiple times and get very, very different responses. A spike could be 33000 counts one run and 8000 the next. I have been in correspondence with Gerstel about this. They are very nice and helpful, but honestly did not know what was going on (neither did I). So, right now I'm not using cryocooling. The inlet just sits at 275C and the desorption flow runs right onto the column. My chromatography is not as good, but it's more reliable.
Feel free to skip over this CIS issue

I don't want to drag anyone else into this craziness, haha.
Thank you very much for your input.
EDIT: If anyone does want to tackle the CIS issues, I suppose the CIS method information would be relevant (this is a separate method than the one I'm currently using and already described)
Initial Temp: -120C
Equilibration Time: .5 min
Initial TIme: .2 min
Ramp:
Rate 12.0C/sec
End Temp: 275C
Hold Time: 3.00 min