has anyone run into analyte degradation in the inlet of your GC? Had the method optimized for a quick runtime, but there were a few caveats and the SOP was getting a little difficult so i went back to the method suggested oven program and gained MCAA and lost DCAA... Any ideas?
This happens many times. With pesticides, carbamates, flame retardants..

If decomposition happens in teh LINER you need to eliminate Glass wool (take a double gooseneck or a cyclosplitter), Use LOWER injection port temperatures (using splitless injection you can us eMUCH lower temperatures as you inject during long time), or use PTV or even better: cold -on-column.

If decomposition happens in the column:(you can see by peak-shape): Make sure we minimize elution temperatures by:
Using Highest flow rate , Flow programming, columns with THIN films, SHORT length and preferrable H2 as carrier gas.
For instance for the BFR's we developed a special "fast" columns based on EPA1614.

rgsd
jaap de zeeuw, Restek corporation
2 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry