leakage in the transducer region
Posted: Sat Mar 31, 2012 4:05 pm
I have an Waters autosampler 717 plus and for some reason the inlet and outlet of the transducer has been leaking slightly, presumably due to higher then usual pressures in that region.
time line
-was getting low areas for my analyte peaks
-checked compressibility test, which passed
-checked injection accuracy test, which did not pass
-called Waters technician who replaced the injection valve, he claims the accuracy test passed, but I checked it afterwards and got 47 uL for 50 uL injections.
-opened up the autosampler and found leakages coming from the inlet to the injection valve, so tightened the screw which ceased the leaking.
-a short while after, found leakages from both the inlet and outlet of the transducer.
-peak areas are getting progressively lower, it's gone to about 1/10 of what it should be. Tightened the bolts on the transducer, still a little bit of water is coming out of the inlet and outlets so pressure in that region must be quite high and increasing.
I unscrewed the inlets and outlets starting from the transducer and found that flow was good from the transducer to the sample loop outlet which connects to the needle. My only guess is that the needle is not properly aligned into the stream and that the port may be sealed up perpetually, which is causing pressure to build in the sample loop/transducer region after successive drawing of sample into the loop.
How often does one need to do a seal pack adjustment? It hasn't been long since I did it but then again one of the sensor bars on the needle platform has its plastic broken, exposing the wires inside of it. I asked my Waters technician about this and he said that it shouldn't be a big deal since the wires do the sensing and the plastic is just the casing. So mmy question to you all is - is it essential that I get the sensor bar fixed? I will get it fixed eventually however.
What measures should I take here? Btw, had all three valves replaced just recently, with the most recent one being the injection valve.
Perhaps the broken sensor bar is causing the seal adjustment to go out of whack every once in a while, which is increasing pressure in the sample loop lines, which in turn is damaging the valves?
time line
-was getting low areas for my analyte peaks
-checked compressibility test, which passed
-checked injection accuracy test, which did not pass
-called Waters technician who replaced the injection valve, he claims the accuracy test passed, but I checked it afterwards and got 47 uL for 50 uL injections.
-opened up the autosampler and found leakages coming from the inlet to the injection valve, so tightened the screw which ceased the leaking.
-a short while after, found leakages from both the inlet and outlet of the transducer.
-peak areas are getting progressively lower, it's gone to about 1/10 of what it should be. Tightened the bolts on the transducer, still a little bit of water is coming out of the inlet and outlets so pressure in that region must be quite high and increasing.
I unscrewed the inlets and outlets starting from the transducer and found that flow was good from the transducer to the sample loop outlet which connects to the needle. My only guess is that the needle is not properly aligned into the stream and that the port may be sealed up perpetually, which is causing pressure to build in the sample loop/transducer region after successive drawing of sample into the loop.
How often does one need to do a seal pack adjustment? It hasn't been long since I did it but then again one of the sensor bars on the needle platform has its plastic broken, exposing the wires inside of it. I asked my Waters technician about this and he said that it shouldn't be a big deal since the wires do the sensing and the plastic is just the casing. So mmy question to you all is - is it essential that I get the sensor bar fixed? I will get it fixed eventually however.
What measures should I take here? Btw, had all three valves replaced just recently, with the most recent one being the injection valve.
Perhaps the broken sensor bar is causing the seal adjustment to go out of whack every once in a while, which is increasing pressure in the sample loop lines, which in turn is damaging the valves?