- Posts: 6
- Joined: Wed Feb 22, 2012 3:34 am
Here is the problem from our waters 2767 autosampler:
Long story short: By injecting a dye sample (with deep blue color), I was able to locate the problem.
I first disconnected the inlet to the column and collected the solvent with a glass bottle. Then I injected 10 ul of the dye sample to the analytical injector (20 ul loop, mobile phase: 100% methanol, 1 ml/min). I didn't get any dye sample collected in the glass bottle. I could see that the dye sample went directly to the waste line. After consulting with the waters tech support, I did the injector rebuilt (replaced the stator face assembly and the rotor seal). But the problem isn't solved. Then I thought maybe there was a blockage in the stator face. I checked the stator face and no blockage found. Plus the system pressure is normal. There is no leaking from the injector either.
The injection to the prep injector (there are two injectors in the 2767 autosampler) works fine, which excludes problems from the probe. Any suggestions or comments? What caused the problem? The waters support suggested me to replace the whole injector valve, which costs ~$3000.
Thanks very much your time and help.