HPLC-IRD Problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
hello, I am having an issue with a HPLC-IRD. After purging, balancing etc the base line sharply decreases on the first injection, then shows a peak which I assume to be a solvent front followed by a decrease in the baseline again. This only happens for the first injection and is fine when stopped, re-balanced and injected again. I have checked the needle, needleseat etc and made new solvents. In addition I have ruled out a problem with the detector as it only happens on the first injection. The method has not been changed and this only started recently. I have a feeling I am overlooking something. Any ideas?
Hello, k.hopkins!

I would suggest the first step is, check out this event on other previously created methods. My first thought - some mystery in the method of the program. Somehow, RID switched off at the first injects. Check computer for a virus, but it is unlikely, i think. (they are also severely may affect the operation of the program).

Chromatograph with a running, turn the valve on the autosampler on the bypass work (without needle) if there is any effect from it? If so, then the problem may be in the valve. Something to sit down them.

This is my first wash and ideas.
Please sorry, if they are too simple.

With Best Regards!
Dmitry
When the path is over and passed successfully, it's nice to remember even the mistakes ...
what model of HPLC are you using?
A. Carl Sanchez
I am using a Shimadzu Prominence HPLC. I have also just noticed i said it is an IRD, its not. I should have written that it is a refractive index detector.
The reason I asked for the model is some instruments have some sort of loop offline option to reduce gradient delay. In other words the loop is removed from the flow path (injection valve placed back in load position) after the injection is made. Under the "right" circumstances this can lead to the contents of loop having a different composition than the mobile phase. RID is especially sensitive to this.
A. Carl Sanchez
thanks for that.Ill check it out on Monday.the advice is appreciated
6 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry