UHPLC method using high pH buffer (pH 10)
Posted: Thu Mar 29, 2012 2:53 pm
Hi, we are running an UHPLC method on Agilent 1290 LC. We are using pH 10 ammonium acetate buffer and max pressure is 800 bar. We had a lot of problem with rotor seals, both on the autosampler and purge valve. Looks like corrosion to me (channels on the seal are completely rusted). We also had pump reliability issues (leaking from pump seal). We had replaced entire pump head assembly twice within a year. We were told by Agilent that we are running out of their specified pH range for 1290 LC. Anyone has any experience with high pressure high pH method on Waters or Dionex UPLC systems? pH 10 is not that high to me....