How can I improve the signal to noise ration for UV detector in HPLC system?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Two ways: more signal, or less noise
More signal:
- inject a bigger sample
- use a stronger mobile phase to get your peaks of interest earlier (narrower) in the chromatogram (isocratic)
- use a steeper gradient to make your peaks narrower
- make sure you detect at an absorbance maximum wavelength for your analyte
- use a smaller column diameter (less dilution of the sample during chromatography)
- get a column of higher efficiency (plate number) to give you narrower peaks.
Less noise:
- if your lamp is old, replace it
- make sure your lamp is properly aligned
- make sure your detector cell is clean. Check the cell windows; if they have "solarized", replace them
- make sure your mobile phase is well degassed
- if you are working at short wavelength (near the UV cutoff of your mobile phase), make sure you are using high-purity solvents
- if you are running a gradient, do the "three-blank-gradient" test to find out if contaminated "A" solvent is contributing garbage peaks.
- Re-run the manufacturer's recommended performance qualification test to verify that the detector is, in fact, operating to its specifications.
More signal:
- inject a bigger sample
- use a stronger mobile phase to get your peaks of interest earlier (narrower) in the chromatogram (isocratic)
- use a steeper gradient to make your peaks narrower
- make sure you detect at an absorbance maximum wavelength for your analyte
- use a smaller column diameter (less dilution of the sample during chromatography)
- get a column of higher efficiency (plate number) to give you narrower peaks.
Less noise:
- if your lamp is old, replace it
- make sure your lamp is properly aligned
- make sure your detector cell is clean. Check the cell windows; if they have "solarized", replace them
- make sure your mobile phase is well degassed
- if you are working at short wavelength (near the UV cutoff of your mobile phase), make sure you are using high-purity solvents
- if you are running a gradient, do the "three-blank-gradient" test to find out if contaminated "A" solvent is contributing garbage peaks.
- Re-run the manufacturer's recommended performance qualification test to verify that the detector is, in fact, operating to its specifications.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 418
- Joined: Mon Aug 15, 2005 3:06 pm
One more: Reduce your dead volume as much as is practical.
Easy and cheap: Minimize the length and diameter of the tubing from the end of your column to your detector.
Less easy, still pretty cheap: Re-plumb your system with narrower bore tubing and be absolutely positive you don't have any leaks. Also make sure that all of your compression fittings are properly made - no ferrules way out at the end of the tubing. Remove any unnecessary unions, too; I've seen patches in the strangest places.
Cheers!
Easy and cheap: Minimize the length and diameter of the tubing from the end of your column to your detector.
Less easy, still pretty cheap: Re-plumb your system with narrower bore tubing and be absolutely positive you don't have any leaks. Also make sure that all of your compression fittings are properly made - no ferrules way out at the end of the tubing. Remove any unnecessary unions, too; I've seen patches in the strangest places.
Cheers!
http://the-ghetto-chromatographer.blogspot.com/
3 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science