TCD or FID for methane gas analysis
Posted: Sun Mar 25, 2012 4:53 pm
by nelson18
How do i analyse methane content of biogas using GC?
Should i use TCD or FID for methane gas.
If it is TCD what should be the carrier gas?
What should be the temperature in injector detector and oven while using TCD?
If it is FID what should be the temperatures?
Can anyone guide me?
What is the retention time of methane in GC?
Nelson
Re: TCD or FID for methane gas analysis
Posted: Mon Mar 26, 2012 6:22 am
by chromatographer1
How do i analyse methane content of biogas using GC?
fixed loop sampling and packed columns are better
Should i use TCD or FID for methane gas.
either is possible. FID is more specific and easier
If it is TCD what should be the carrier gas?
Helium, hydrogen, or argon
What should be the temperature in injector detector and oven while using TCD?
I would use minimum of 80C for injector, 150C for detector, and 60C for oven
If it is FID what should be the temperatures?
same
Can anyone guide me?
maybe
What is the retention time of methane in GC?
varies by column length, type, packing, coating, and linear flow rate
Re: TCD or FID for methane gas analysis
Posted: Mon Mar 26, 2012 10:35 am
by AICMM
nelson18,
If it were me and all I had to do was methane, I would second Rod's view and go with an FID. Then methane will be a nice simple peak early in the chromatogram. The other advantage is going with an FID is you can measure high levels or also low levels easier than with a TCD.
The unspoken question is.... do you really want to measure all the other stuff (like lots of other posters) as well? There is likely to be ethane, CO, CO2, possibly H2, etc.... depending on your process and this gets much more complicated.
Best regards,
AICMM