Agilent 7694 Installation into Chemstation
Posted: Sat Mar 24, 2012 12:23 am
by ChromStud5.0
Any guru's out there know how to install the Headspace Sampler to Chemstation? The error message is "cannot find chemstation". Chemstation is B.03.01 and the headspace is running A.01.06. I have them both installed on another machine that works fine but can't seem to connect with this configuration. Thanks for any help!!
Re: Agilent 7694 Installation into Chemstation
Posted: Sat Mar 24, 2012 2:36 pm
by aldehyde
Do you see the headspace info in your instrument menu when you open chemstation? there is a configuration page where you provide it with the headspace's IP address and everything communication wise. You may also have to hit the reconnect button to refresh the chemstation-headspace communication.
If you have the headspace and GC plugged into a hub you should try pinging the headspace and make sure you can communicate on that level before involving chemstation.