Changing tungsten lamps in perkin elmer series 200 DAD

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Any tips?

Do you need UV safety specs?
1) the lamp should be off when you go to change it, so I'd not worry about UV specs.

2) Make sure the lamp's bulb is very clean and that you leave no fingerprints on it.

3) Find your detector's specs for sample and/or reference energy and make sure that they are met (and that this represents an improvement over readings from the old lamp) after the new lamp is installed.
Thanks,
DR
Image
1) the lamp should be off when you go to change it, so I'd not worry about UV specs.

2) Make sure the lamp's bulb is very clean and that you leave no fingerprints on it.

3) Find your detector's specs for sample and/or reference energy and make sure that they are met (and that this represents an improvement over readings from the old lamp) after the new lamp is installed.

Thanks, but this is a perkin elmer series 200 DAD using total crap, I mean total chrom. I dont think there is anywhere to check lamp strenght. Unless someone can enlighten me?
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 2 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry