I've been tasked with devising a GC protocol for a relatively challenging application. The samples we analyze come from the steam cracking of petroleum which are then typically subjected to some form(s) of treatment afterward, such as distillation or heat soaking. In the streams we look at, the major component is DCPD (dicyclopentadiene), which undergoes a retro Diels-Alder at 150C forming CPD (cyclopentadiene), and therein lies our problem.
Because we have temperature sensitive analytes, we hold our inlet at 160C (6890, Rtx-5 column, split mode) which is definitely not hot enough. However, after heat-soaking the material, we get trimer, tetramer, and possibly higher MW oligomers that all boil well beyond 160C. You can probably see where this is going...
Our problems:
1) Can't heat inlet too high because of temperature sensitive analytes, but need to vaporize high MW analytes--problems with discrimination and inlet contamination
2) Can't run the GC too hot or we see rearrangement products, but need to in order to cause elution of high MW analytes
3) Residual C5's in sample may be lost to backflashing
What I've done so far is affirmed that the GC method used to analyze these samples (set up before I began) is inadequate by running a variety of tests using trimer/tetramer standards, solvent blanks, column bakes, and inlet bakes. By various methods, I also confirmed that our contamination source is the inlet. With a slight pressure ramp and some tweaking, I've also come up with a method that elutes our higher boilers in a reasonable amount of time.
I think the obvious step is to try running in splitless mode to address a large portion of the difficulties we are having. I'm hoping that running in splitless mode with perhaps a double-tapered liner will be the silver bullet for this application as it will address backflashing, injector discrimination issues, and possibly even the inlet contamination we've been seeing (because of the residence time of the samplein the inlet, possinly vaporizing the high boilers).
My question to the community here is what are our other options?
Because we have temperature sensitive analytes, we hold our inlet at 160C (6890, Rtx-5 column, split mode) which is definitely not hot enough. However, after heat-soaking the material, we get trimer, tetramer, and possibly higher MW oligomers that all boil well beyond 160C. You can probably see where this is going...
Our problems:
1) Can't heat inlet too high because of temperature sensitive analytes, but need to vaporize high MW analytes--problems with discrimination and inlet contamination
2) Can't run the GC too hot or we see rearrangement products, but need to in order to cause elution of high MW analytes
3) Residual C5's in sample may be lost to backflashing
What I've done so far is affirmed that the GC method used to analyze these samples (set up before I began) is inadequate by running a variety of tests using trimer/tetramer standards, solvent blanks, column bakes, and inlet bakes. By various methods, I also confirmed that our contamination source is the inlet. With a slight pressure ramp and some tweaking, I've also come up with a method that elutes our higher boilers in a reasonable amount of time.
I think the obvious step is to try running in splitless mode to address a large portion of the difficulties we are having. I'm hoping that running in splitless mode with perhaps a double-tapered liner will be the silver bullet for this application as it will address backflashing, injector discrimination issues, and possibly even the inlet contamination we've been seeing (because of the residence time of the samplein the inlet, possinly vaporizing the high boilers).
My question to the community here is what are our other options?