repetitive peaks in gradient of blank run
Posted: Sun Mar 18, 2012 12:05 am
I am running an impurity test for a polypeptide in a C18 column and using 226nm detection. My problem is that repetitive peaks appear at retention times that overlap the impurity I'm supposed to test. These peaks are a set of tree repetitions of 4 peaks each, always at the same retention times. These peaks appear even when running blanks.
I'd really appreciate any suggestion to solve this problem. Below I describe some details of the run and the solutions I've tried. thanks a lot!
Peaks grow with every injection until reaching a maximun size after 20 injections approx. Peaks appear between times 30 and 40 minutes.
I could not eliminate these peaks by:
-washing the column
-replacing every component of mobile phases
-running the blank in 3 different hplc (always the same peaks are there...)
Gradient is composed of mobil phase A: H2O 93%/ACN 7% with TFA 0.13%, and mobile phase B is: H2O 30%/ACN 70% with 0.1% TFA.
Gradien is as follows:
0-30%B
25-55%B
40-100%B
45-100%B
50-30%B
55-stop
I'd really appreciate any suggestion to solve this problem. Below I describe some details of the run and the solutions I've tried. thanks a lot!
Peaks grow with every injection until reaching a maximun size after 20 injections approx. Peaks appear between times 30 and 40 minutes.
I could not eliminate these peaks by:
-washing the column
-replacing every component of mobile phases
-running the blank in 3 different hplc (always the same peaks are there...)
Gradient is composed of mobil phase A: H2O 93%/ACN 7% with TFA 0.13%, and mobile phase B is: H2O 30%/ACN 70% with 0.1% TFA.
Gradien is as follows:
0-30%B
25-55%B
40-100%B
45-100%B
50-30%B
55-stop