Thank you for the reply. However, it is definitely not stereoisomers, since the molecule has no stereocenters or potential rotomers or cis/trans olefins. It is a very simple bis-aryl ether. However, I see two peaks by GC. Some concentrations worse than others. I used isopropanol as the solvent to do serial dilutions, which gave two peaks, and then I tried acetonitrile, which gave two peaks sometimes, but was much better. From my search of the forums, it seems someone else noted splitting of peaks (see original post for reference). I just don't know how to explain this.The most simple observation is the possibility that you have a stereoisomer that is not easily distinguished by H1 NMR. The GC was able to seperate the 2 isomers.
You originally asked about integration and calibration, which suggests that you need quantitative results. The conditions that you give are not likely to yield good repeatability. In particular a 5 ul aqueous injection, even if split, is likely to cause all sorts of trouble. Am I right to suspect that separating your target compound from potential interferences is not a problem ? If so there are probably better ways of solving the peak splitting. What else did you try ?Thank you for the information. I was able to solve the problem simply by taking the compound up in ethanol and then diluting with water to give a mostly aqueous solution for injection. Sorry I cannot disclose the compound I am working with, but it is a bis-aryl ether, which is not very soluble in water, but seemed to give stable solutions at low concentrations. As a matter of reference, the conditions I used were 175 degrees to 210 degrees over a 6 min time interval, with a column flow rate of 1.03 mL/min, on a 30 m Restek RTX-XLB capillary column (a proprietary non-polar phase), with an autoinjector, with injection temperature set at 210 degrees C, 5 microL injection, and the compound came out at 4.5 min.
Thanks for your help.
Don's advice is good. Another thing that you can try is to add salt to the AcN/water mix - the AcN will separate and although AcN is not commonly used as a solvent in GC it will be much less susceptible to strange inlet effects than a 1:1 mix with water.On the low end I am trying to measure approximately 100 nM solution. The ether is soluble in most any organic solvent such as DMSO, diethyl ether, chloroform, methylene chloride, methanol, acetonitrile, etc. The clean-up is fairly simple, an 100 microL aliquot is removed from the medium and added to 100 microL of ice cold acetonitrile. The solution is then centrifuged and the supernatant was used directly for GC-MS analysis. I said earlier it was mostly aqueous, but I guess it is actually 50% aqueous.
Right, the salting out is the basis of the QuEChERS extraction method that I'll be working with in my lab for pesticides - MgSO4 and NaCl are shaken with the mix. Apparently salting out also works with isopropanol. The ACN layer after salting out and centrifuging is approximately 16% water, which can be further diminished by a second MgSO4 treatment if desired.Don's advice is good. Another thing that you can try is to add salt to the AcN/water mix - the AcN will separate and although AcN is not commonly used as a solvent in GC it will be much less susceptible to strange inlet effects than a 1:1 mix with water.
A back of envelope calculation gives you about 15 ng per ul in what you inject to the GC - I'm guessing a MW of 300 for your analyte. If your MS is working right you could do a 1 ul injection with a 10:1 split and still have a very respectable signal:noise ratio at the lower end of your calibration.