Combi-PAL Liquid and HEadspace questions and Liners
Posted: Fri Mar 16, 2012 3:41 pm
I have never used a Combi-PAL but have one that has Liquid and headspace injection options on a 7890 witha an FID... and I have some questions.
The standard Comb-PAL liquid syringe is 10 uL.
In the manual it says that syringe should be used for 1-10 uL... But the overview card lists sample volumes for that syringe down to 0.2uL... Can you get good reproducible injections in the 0.4-0.5uL range with that syringe?
I see there are 5 and 1.2uL syringes available...If I can;t use the 10 uL would the 5 uL be OK or would I need the 1.2 uL?
Does each size need it's own syringe holder plate? If so changing syringe sizes (at least the 1st time) gets more expensive than then just buying another syringe!
Now headspace. I have only done sample loop headspace with the Agilent sampler before not a syringe based unit. On the Agilent HS sampler I use the deactivated 2mm ID straight through liner. It occurs to me as I am injecting gas (a mL on the Agilent HS) into the inlet that a liner with a larger internal volume might be better (I know injection rate matters - some analytes are likely to be migrating at the staring temp so I don't want to have to inject too slowly).
Is that true? Any specific liner recommendations?
Thanks,
- Karen
The standard Comb-PAL liquid syringe is 10 uL.
In the manual it says that syringe should be used for 1-10 uL... But the overview card lists sample volumes for that syringe down to 0.2uL... Can you get good reproducible injections in the 0.4-0.5uL range with that syringe?
I see there are 5 and 1.2uL syringes available...If I can;t use the 10 uL would the 5 uL be OK or would I need the 1.2 uL?
Does each size need it's own syringe holder plate? If so changing syringe sizes (at least the 1st time) gets more expensive than then just buying another syringe!
Now headspace. I have only done sample loop headspace with the Agilent sampler before not a syringe based unit. On the Agilent HS sampler I use the deactivated 2mm ID straight through liner. It occurs to me as I am injecting gas (a mL on the Agilent HS) into the inlet that a liner with a larger internal volume might be better (I know injection rate matters - some analytes are likely to be migrating at the staring temp so I don't want to have to inject too slowly).
Is that true? Any specific liner recommendations?
Thanks,
- Karen