A question about recovery of solid phase extraction

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

8 posts Page 1 of 1
Dear bro,

May I ask you a question about the recovery of SPE ?

I use C18 cartridge for lindane (organochloride pesticide) extraction from water sample.

I validate the method to test the recovery of method with concentrations of 1.5, 0.5, 0.1 mg/l . The natural pH is about 5 - 6.

But the recovery is very high : 115 ~ 125 %.

According to AOAC standard, the recovery with this range of concentration should be about 80 -110%.

I would like to ask the reason why the recovery is great like that ? And why the recovery is higher than 100%.

Thank you so much !
You are probably seeing some matrix enhancement. Are you running on GCMS or GC/ECD?
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
Have you run any blanks ?, do they have peaks at lindane's retention time ?

Peter
Peter Apps
You are probably seeing some matrix enhancement. Are you running on GCMS or GC/ECD?


I'm running GC/ECD

Have you run any blanks ?, do they have peaks at lindane's retention time ?

Peter
I ran the blank sample already. There is a peak at lindane's retention time.

Thanks for your reply !
I think Peter has discovered the problem. Halogenated compounds are ubiquitous in the environment. Your contamination is adding to your peak height/area.
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
I think Peter has discovered the problem. Halogenated compounds are ubiquitous in the environment. Your contamination is adding to your peak height/area.
But my sample is simulated sample, these samples are not from environment.
No matter where the samples and reagents come from, if you have a peak in the blank it will add to the lindane peak in recovery tests, making it bigger than it would be if the peak was due only to the quantity of analyte that you spiked for the recovery. The bigger peak will lead to a result for a higher quantity of lindane, which will give you spuriously high recoveries. The preferred solution is to operate clean enough to reduce the interfering peak to neglible levels, or to change chromatographic conditions to separate the lindane from the interference. Second choice is to subtract the blank peak area from the lindane peak area, but you can only do this if the interference is at the same level all the time.

Peter
Peter Apps
No matter where the samples and reagents come from, if you have a peak in the blank it will add to the lindane peak in recovery tests, making it bigger than it would be if the peak was due only to the quantity of analyte that you spiked for the recovery. The bigger peak will lead to a result for a higher quantity of lindane, which will give you spuriously high recoveries. The preferred solution is to operate clean enough to reduce the interfering peak to neglible levels, or to change chromatographic conditions to separate the lindane from the interference. Second choice is to subtract the blank peak area from the lindane peak area, but you can only do this if the interference is at the same level all the time.

Peter
Many thanks for your advice !
8 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry