- Posts: 64
- Joined: Mon Dec 05, 2011 11:41 pm
I'm working a Gerstel TDSA2 connected to an Agilent 7890 with a microECD detector. I'm looking at Chloroform, 1,1,1-Trichloroethane, and Carbon Tetrachloride. We are collecting on Gerstel Carbotrap 300 tubes. I have been working on optimizing the method on the GC, to good effect. However, I know very little about the thermal desorption. I'm not sure the parameters are the best. I suspect I'm not getting the compounds on the column as quickly/cleanly as possible, and such the peaks may be wider. My current parameters are as follows:
_______________________________
Initial:
Initial Temp: 50C (Actual 65C)*
Delay Time: 0.20 min
Initial Time: 1.00 min
Ramp:
60C/min to 275C, hold 5 min
Settings:
Transfer Temp: 275C (Actual 275C)
Vent Time: 1.00 min
Desorption Mode: Solvent Vent
Sample Mode: Sample Remove
I don't know what the gas flow is. The physical controller hooked up to the Gerstel stuff says "Adsorption flow: 21 ml/min", but the software says "Adsorption flow: 50 ml/min".
I'm splitting the flow at my inlet at about 20:1. We look at low levels of chloroform (which is not the best ECD-sensitive compound), so it necessitates the lower split ratio.
_______________________________
*What does this mean? Why is the "Actual" temp different?
Does anyone have experience in this area? I would appreciate some pointers on what I should try. I don't particularly understand the relationship between all the parameters. I'm having a hard time finding anything online, and I don't have any useful material from the manufacturer.
Thanks.