ASTM Noise Test

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I just changed out my deuterium lamp on an Agilent 1100 series HPLC. Right before I changed this lamp, I did the DAD self-test, and discovered failing results from the 20 minutes ASTM noise test. Now, I went to do the DAD self-test again, but now once it runs through the self-test (which are all passing), it just says the self test is finished without running the very last test, the 20 min. ASTM noise test. What am I missing? I do have it checked in the dialog box as well, so I don't understand why it is doing this. :idea:
ıf you change a deuterıum lamp, you would wait to lamp energy about 1 hour. When ı changed deuterıum lamp ı did a lamp calibration automatically in chemstation software.

regards
Did you accidentally press "Skip all remaining tests" just before the last one?
Q. F. Ignacio Viera
3 posts Page 1 of 1

Who is online

In total there are 377 users online :: 1 registered, 0 hidden and 376 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 376 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry