Phosphate buffers with different pH values and 0.1%TFA are normally used reagents in mobile phase to improve peaks' separations and shapes.
In general, could any expert explain when we use which ?
Thanks.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
TFA is not good IP reagent, you need at least C4 carbon chain to have any IP properties. People call TFA ion-pairing reagent but it is a "urban myth" at least for small molecules. If you take a basic molecule and RP column you are not going to see much improvement in retention time by using TFA vs. phosphoric acid. The only difference might be is poor peak shape in case of phosphoric acid due to residual silanol interactions (because you create lower pH with TFA assuming same concentrations)
In my opinion TFA makes peptides and proteins more polar a it protonates amino groups, in some cases you might see slight increase in hydrophobicity if peptide has a lot of free carboxylic acids and you suppress ionization of these acids and make it less polar.TFA is not good IP reagent, you need at least C4 carbon chain to have any IP properties. People call TFA ion-pairing reagent but it is a "urban myth" at least for small molecules. If you take a basic molecule and RP column you are not going to see much improvement in retention time by using TFA vs. phosphoric acid. The only difference might be is poor peak shape in case of phosphoric acid due to residual silanol interactions (because you create lower pH with TFA assuming same concentrations)
Thanks fro the reply.
I think TFA as an IP reagent helps to improve separation and peak shape especially for peptides and proteins. These components obviousely possess long carbon chains. The machanism is that the trifluoroacitates associate with the positively charged groups on peptite bonds leading to the polarity reduction of peptide/protein molecules, and resulting in better intereation with RP column stationary phase. Small molecules don't show much improvements are probably due to not having too many positively charged sites on target molecules. Is this correct?
Yes, all depend on the actual peptidre/protein targeted. The proton will combine with NH2-end group to for NH3+, while COOH remaining on neutral if an acidic enviroment is provided, thus the negative acetate, CF3COO- associates with NH3+ end of the target molecule to become an ion-pair. Therefore in overall its hydrophobicity indreased because of the ion-pair. Is this explanation reasonable?In my opinion TFA makes peptides and proteins more polar a it protonates amino groups, in some cases you might see slight increase in hydrophobicity if peptide has a lot of free carboxylic acids and you suppress ionization of these acids and make it less polar.TFA is not good IP reagent, you need at least C4 carbon chain to have any IP properties. People call TFA ion-pairing reagent but it is a "urban myth" at least for small molecules. If you take a basic molecule and RP column you are not going to see much improvement in retention time by using TFA vs. phosphoric acid. The only difference might be is poor peak shape in case of phosphoric acid due to residual silanol interactions (because you create lower pH with TFA assuming same concentrations)
Thanks fro the reply.
I think TFA as an IP reagent helps to improve separation and peak shape especially for peptides and proteins. These components obviousely possess long carbon chains. The machanism is that the trifluoroacitates associate with the positively charged groups on peptite bonds leading to the polarity reduction of peptide/protein molecules, and resulting in better intereation with RP column stationary phase. Small molecules don't show much improvements are probably due to not having too many positively charged sites on target molecules. Is this correct?
Hi VladIn you case I would try to stay with isocratic conditions. When you have compounds which are very similar you want to stay with isocratic. Your another alternative would be chiral chromatograpphy but it is more expensive. If are are willing to send me samples we can try to develop method for you free of charge on one of our mixed-mode columns. You can also try another IP reagent with longer chain.
Two questions:In you case I would try to stay with isocratic conditions. When you have compounds which are very similar you want to stay with isocratic. Your another alternative would be chiral chromatograpphy but it is more expensive. If are are willing to send me samples we can try to develop method for you free of charge on one of our mixed-mode columns. You can also try another IP reagent with longer chain.
1. The calculated organic con for isocratic elution is alout 72%.Two questions:In you case I would try to stay with isocratic conditions. When you have compounds which are very similar you want to stay with isocratic. Your another alternative would be chiral chromatograpphy but it is more expensive. If are are willing to send me samples we can try to develop method for you free of charge on one of our mixed-mode columns. You can also try another IP reagent with longer chain.
1. For a gradient example, the organic phase was 25-95% running 60 min with 0.8mL/min flow rate, and the peak eluted at 40min, how should I estiamte the organic concentration if ruunning isocratic in stead?
2. When running gradient, the IP should be put in the water or organic phase, or both phases?
1. 72%. Perfect!1. The calculated organic con for isocratic elution is alout 72%.Two questions:In you case I would try to stay with isocratic conditions. When you have compounds which are very similar you want to stay with isocratic. Your another alternative would be chiral chromatograpphy but it is more expensive. If are are willing to send me samples we can try to develop method for you free of charge on one of our mixed-mode columns. You can also try another IP reagent with longer chain.
1. For a gradient example, the organic phase was 25-95% running 60 min with 0.8mL/min flow rate, and the peak eluted at 40min, how should I estiamte the organic concentration if ruunning isocratic in stead?
2. When running gradient, the IP should be put in the water or organic phase, or both phases?
2. The IP should only be added to the water as dissolve issues with organic phase.
3. In addition, why not give a try to ace? As in some cases, ace is complementary with met.