Schematics for Varian 3400
Posted: Tue Mar 13, 2012 7:49 pm
by Redi
I started browsing the forum and it surprised me when I found so much users for Varian 3400. I´m having problem with one of these (power suply overheating) and need the schematics to check some conections. Anyone knows where can I find these schematics? Or service manual?
Redi Siqueira
Re: Schematics for Varian 3400
Posted: Fri Apr 20, 2012 12:10 am
by HbJ
Please describe the exact nature of your problem :)
Posted: Fri Apr 20, 2012 11:59 am
by Redi
The cable that goes from power suply to J31 on the mainboard has overheated. I replaced the wire and took the voltage reading (20v) but I don´t know if this is the right. The GC started ok (but the error messages I mencioned at the other post).
Till now, it is working well, but I would like to have a better understanding of the equipment.
Thank you for your interest.
Redi
Posted: Fri Apr 20, 2012 9:35 pm
by HbJ
There is no J31 in any schematics I have.
Please re-check and take a picture so I can see the actual position.
It might be J21 which carries the net voltage.
Posted: Tue Apr 24, 2012 1:44 pm
by Redi
I sending you a photo of mainboard and signed J31 conector and also a little burned path.
First time I try to put an image here, if you can not see it, please, let me know.
Thanxs,
Posted: Tue Apr 24, 2012 4:26 pm
by HbJ
I don't see an image, please give the URL.
Posted: Tue Apr 24, 2012 4:44 pm
by Redi
Posted: Tue Apr 24, 2012 11:05 pm
by HbJ
Well... the part number hints towards a mainboard for a Varian 3600.
The burned path carries one of the terminals of the ignitor voltage for the FID (J 18).
None of my diagrams show a J31 but the visible connected line between J 18 and J 31 on the circuit board carries one terminal of the voltage for TSD no. 2.
Posted: Wed Apr 25, 2012 12:05 pm
by Redi
HbJ:
Do you know which would be the rigth voltage for the ignitors? Both of them (2 FIDs in the system) are open and I´m thinking if its not related with this conector.
Posted: Wed Apr 25, 2012 4:19 pm
by HbJ
Open in what way?
If you can't ignite them, it's said connector. According to the specs, it's delivering 3.8 V to the ignitor coils.
Posted: Wed Apr 25, 2012 6:29 pm
by Redi
The coils are open. The analyst is doing the ignition manually, with an eletric lighter till new ones arrive, but I´m afraid the new ones can burn again if the circuit is not ok. I even thinking in recommend they do not change the ignitor probe and live with the problem till buy a nem GC.
At J31 conector (wires from power suply to conector) I read 20 v !!
Posted: Wed Apr 25, 2012 9:11 pm
by HbJ
That is okay as J31 is the TSD voltage which is around 18 volts.
J18 carries the ignitor voltage.
Posted: Mon May 14, 2012 2:33 am
by HbJ
Just a hint: The ignitor voltage must only be present when ignition is demanded. If it's stuck on you have an electrical problem (just saying).
Posted: Mon May 14, 2012 3:46 pm
by Redi
OK, HbJ,
The system is running without the ignition (I´m using an external ignitor) and seems is ok. I´m still afraid to replace the ignitor for a new one and burn it again.
Thanxs a lot.
Posted: Tue May 15, 2012 10:03 pm
by HbJ
OK, HbJ,
Thanxs a lot.
That's fair :) Have fun with your 3400!