We have a dirty flow cell on our Cap LC and cannot get it clean. We have tried 100% H2O and 100% MeOH (overnight) but the back pressure is still very high.

Can anyone give some suggestions of what to try next?.

For info - we have been running biological samples in ammonium formate/acetate + MeCN / MeOH mobile phase.

Thanks all!.

James

Posted: Thu May 12, 2005 12:52 pm
by JM
Try forcing dilute nitric acid in the flow cell by manual glass syringe.

JM

Posted: Fri May 13, 2005 8:24 am
by HW Mueller
If there is a possibility of having proteins, etc., in there you may want to use detergents, maybe with dithiothreitol, or urea (or other chaotiopes) first.

thanks

Posted: Fri May 13, 2005 2:05 pm
by tayzyboy
Thanks, I'll pass that on to our Cap-LC analyst

James

Dirty Flow Cell

Posted: Mon May 16, 2005 6:28 pm
by Chris Pohl
tayzyboy,

Did you try reversing the flow to the flow cell? This doesn't always work but often a partial blockage is effectively removed by simply reversing the low for a few minutes and then restoring flow to the normal flow direction.