Hello to everyone,
I want to quantify 14 compounds in urine sample so is it required to use 14 internal standards for each individual compound or i can use only one internal standard,please help.


Thanks
Chandra
It depends what you want to correct for. If you are correcting for cosuppression and have 14 chemicals eluting at different times, you will need 14 internal standards eluting at the same times, because cosuppression is an interaction of two chemicals that happen to coelute. If you are correcting for general loss of instrument efficiency as it gets dirty, a single internal standard reasonably chemically-similar to the compounds will already help a lot. If you are compensating for loss of compound during extraction, your internal standard(s) must be chemically similar to the measured analytes, so you can be convinced that the loss of internal standard will match that of measured analyte. That means that you may need only 1 internal standard if all your compounds are closely related, or more, if they are diverse.
2 posts Page 1 of 1

Who is online

In total there are 192 users online :: 0 registered, 0 hidden and 192 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 192 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry