Mass spec interpretation question
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
In the GC/MS run of certain amines, I see a mass of 70 Daltons greater than the molecular ion (molecular ion itself is negligible). Is it possible and why?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
This does not sound to be right. If you have severly overloaded peaks, one might be able to postulate ion-molecule reactions to give ions greater than the molecular weight of the compunds. But even in that case, one should see the expected spectrum for the compounds at the very beginning and very end of the peak -- where the concentration would be typical of what one sees in a GC/MS run. And, severly overloaded peaks tend to give severly distorted spectra due to overloading of the detector.
If you have severly overloaded peaks, dilute and shoot again to get good spectra. If the peaks are not overloaded, you are most likely looking at something other than what you expected.
If you have severly overloaded peaks, dilute and shoot again to get good spectra. If the peaks are not overloaded, you are most likely looking at something other than what you expected.
-
- Posts: 34
- Joined: Sun Jan 08, 2012 5:37 am
thanks for the reply. The peak is not overloaded but I do beleive I'm seeing something else (despite AMDIS 100% library hit). How about another compund of interest which does have large molecular ion peak (and other ions per NIST are present as well) but there is still even a larger peak 70 Daltons greater than the molecular ion.
This does not sound to be right. If you have severly overloaded peaks, one might be able to postulate ion-molecule reactions to give ions greater than the molecular weight of the compunds. But even in that case, one should see the expected spectrum for the compounds at the very beginning and very end of the peak -- where the concentration would be typical of what one sees in a GC/MS run. And, severly overloaded peaks tend to give severly distorted spectra due to overloading of the detector.
If you have severly overloaded peaks, dilute and shoot again to get good spectra. If the peaks are not overloaded, you are most likely looking at something other than what you expected.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
If these "certain amines" are drugs, and you are using an alcohol to extract them then there is some discussion inthe archives somewhere that might be relevant - apparently they can react with amides to form heavier compounds.
Peter
Peter
Peter Apps
4 posts
Page 1 of 1
Who is online
In total there are 206 users online :: 2 registered, 0 hidden and 204 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot] and 204 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot] and 204 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science