Hi to all can any one explain me how to define Range and Linearity in validation.
Regards
A.Kalidass
Linearity is normally related to a calibration curve, it is a range of concentrations for which the responses can be fitted to a linear function of a certain defined qualities.

Range is the interval of concentrations to which the method is applicable. If the method should give 70-140% recovery and 35% RSD of duplicates, then the range is the interval of concentrations where these characteristics will be routinely achieved all the times, e.g. range - LOD-ULL for Bisphenol A in oil from 10 ppb to 100 ppm by a method “A”.
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
Thank you for the explanation
Regards
A.Kalidass
3 posts Page 1 of 1

Who is online

In total there are 75 users online :: 1 registered, 0 hidden and 74 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 74 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry