To JI2002
I just downloaded the HP FlowCalc 2.0 from the Agilent site (
www.chem.agilent.com/cag/main.html) and played with it. I also called Agilent for clarifications.
The terms "Outlet Flow (mL/min)" and "Inlet Flow (mL/min)" in the FlowCalc mean the following:
"Outlet Flow (mL/min)" is a column volumetric flow rate measured at atmospheric pressure, AtmPress, and at some reference temperature, RefTemp, which is the same for all measurements regardless of actual column temperature. What exactly RefTemp is, is not clear, but it looks like RefTemp = 0C (more about it later).
"Inlet Flow (mL/min)" is the column volumetric flow rate measured at inlet pressure and temperature. (This measure of the column flow is necessary for calculation of time that it takes for the column flow to sweep an inlet liner of a known volume.)
If Outlet Flow, Fout, inlet gauge pressure, pg, and inlet temperature, Tin (in degree Celsius), are known then Inlet Fow, Fin, can be found as:
Fin = ((273 + Tin)/(273 + RefTemp)).(AtmPress/(AtmPress + pg)).Fout
where AtmPress = 1 atm = 14.7 psi.
Your Example:
Column and setpoints:
Column 30m-0.25mm
Oven temperature 180C
Gas helium
Outlet flow, Fout = 1 mL/min
Inlet temperature: Tin = 175C
Calculated parameters of helium flow:
Inlet pressure: pg = 15.1 psig
Average velocity: 38.1 cm/s
Inlet flow, Fin = ((273+175)/(273+0)) ((14.7/(14.7+15.1)) 1 mL/min = 0.809 mL/min
Comment:
It appears from the calculation of Inlet flow in the example (and from several other test of FlowCalc that I made) that the reference temperature, RefTemp, for the measurement of Outlet flow in the FlowCalc is 0C. On the other hand, as far as I know, in Agilent 6890 GC, in Agilent 6850 GC, and in the instruments from other vendors, the column flow rate is assumed to be measured at 25C. Keep that possible discrepancy in mind.