carryover of organic acids by GC
Posted: Fri Mar 02, 2012 2:47 pm
I have been analyzing a series of monocarboxylic acids and naphthenic acids (C11-C20) by GC-FID using a Perkin Elmer Autosystem with Programmable Splitless (large volume 20-30uL) injections.
The solvent is 1% formic acid in methanol. Direct analysis with no dervitization using an Agilent PEG column.
concentration range is ~ 1 ppm - 40 ppm.
I am having signifcant problem with injection-to-injection carryover. Blank GC runs (no solvent injection) are clean.
Checked cleanliness of solvents but they too are clean. Have switched out syringe and inlet liner and that also is not issue.
Also observing "priming" phenomena where I need to do a number of injections of same sample or standard to achive reasonably constant response.