Showing any chromatograms, we know the unit for peak area is "mV/sec". However, what is the unit for peak HIGHT please?

Re: Peak hight

Posted: Thu Mar 08, 2012 9:56 pm
by cawarhur
Guessing you mean peak height...Its pA--picoamps.

Re: Peak hight

Posted: Fri Mar 09, 2012 12:07 am
by wulisheng2
Guessing you mean peak height...Its pA--picoamps.
Ops, you are exactly right that is Peak Hight.

Thank you so much!

Re: Peak hight

Posted: Fri Mar 09, 2012 7:12 am
by Peter Apps
Guessing you mean peak height...Its pA--picoamps.
So if area is in mv.s, how do you get from a height in pA and width in seconds to the area ?

Peter

Re: Peak hight

Posted: Sat Mar 31, 2012 8:27 am
by wulisheng2
Guessing you mean peak height...Its pA--picoamps.
So if area is in mv.s, how do you get from a height in pA and width in seconds to the area ?

Peter

In my practise, I use the mV/sec as the unit of Peak Area versus concetrations for establishing a regression coordinate. The peak areas are obtained from the integration of a peak on chromotogram where the y axis is absorbance unit and x is time. So to perform an integration for peak area needs a range, the time points from peak initial to end. A peak integration result is a measure of area that consists of unlimited numbers of line measures between the two time points. However, Peak Hight does not, as it is a length measuement that is a verdical line from peak top point to baseline. That means no integration needed for Peak Hight.

Re: Peak hight

Posted: Sat Mar 31, 2012 9:34 am
by Peter Apps
Guessing you mean peak height...Its pA--picoamps.
So if area is in mv.s, how do you get from a height in pA and width in seconds to the area ?

Peter

In my practise, I use the mV/sec as the unit of Peak Area versus concetrations for establishing a regression coordinate. The peak areas are obtained from the integration of a peak on chromotogram where the y axis is absorbance unit and x is time. So to perform an integration for peak area needs a range, the time points from peak initial to end. A peak integration result is a measure of area that consists of unlimited numbers of line measures between the two time points. However, Peak Hight does not, as it is a length measuement that is a verdical line from peak top point to baseline. That means no integration needed for Peak Hight.
OK, let me put it another way; to measure an area in mV.s you need linear measurements in mV, not in pA. How do you get from pA to mV ?

Peter

Re: Peak hight

Posted: Mon Apr 02, 2012 6:20 pm
by uzman
Current and voltage can be converted to each other if a series resitor is used.
Remember ohm's law : V= I x R

Re: Peak hight

Posted: Mon Apr 02, 2012 7:08 pm
by GasMan
The question on how one can convert pA to mv is the electronic equivalent of asking on how one can convert flow to pressure. You can not convert flow to pressure without knowing other factors. For a capillary column you need to know the length, internal diameter, temperature and the viscosity of the gas. With this information you could then convert flow to pressure. For a packed column this would not be possible.

As uzman points out, you can convert pA to mv if you know the resistance of the circuit. If you are talking about an amplifier, there is no easy answer, without knowing the circuitry in detail. This means that pA to mv can vary from manufacturer to manufacturer. For the same manufacturer it could change from GC model to GC model, and on the same instrument of a manufacturer it can vary from the digital sigal output and the analogue output.

In summary, there is no general rule to convert pA to mv, other than Ohm's law as already pointed out, which only applies to a simple resistor.

Gasman

Re: Peak hight

Posted: Tue Apr 03, 2012 2:25 am
by wulisheng2
I am apt to accept Gasman's points, we may not get adequate factors to convert PA to mV and I think it requires more complicated math machanisms.

We may pay attention to another observation that is the coordinate of the Voltage/time (uV/min) while a GC is running. The y axis unit is uV. So could we say the unit for Peak Hight is pico or micro Volt(V), rather than Ampere(A)?

Thanks

Re: Peak hight

Posted: Tue Apr 03, 2012 12:23 pm
by JI2002
We may pay attention to another observation that is the coordinate of the Voltage/time (uV/min) while a GC is running. The y axis unit is uV. So could we say the unit for Peak Hight is pico or micro Volt(V), rather than Ampere(A)?

Thanks
That's the point Peter was trying to make.

Re: Peak hight

Posted: Fri Apr 06, 2012 10:49 am
by Peter Apps
We may pay attention to another observation that is the coordinate of the Voltage/time (uV/min) while a GC is running. The y axis unit is uV. So could we say the unit for Peak Hight is pico or micro Volt(V), rather than Ampere(A)?

Thanks
That's the point Peter was trying to make.
Indeed it was !

Peter

Re: Peak hight

Posted: Mon Apr 09, 2012 11:57 am
by jdezeeuw
Now we're on this signal discussion:

why is the Y-signal in MS always listed as "counts"? Whats SI dimension of [counts]?. I wanted to ask that for quite some time being a non-MS specialist.

jaap

Re: Peak hight

Posted: Mon Apr 09, 2012 6:16 pm
by dblux_
Now we're on this signal discussion:

why is the Y-signal in MS always listed as "counts"? Whats SI dimension of [counts]?. I wanted to ask that for quite some time being a non-MS specialist.

jaap
"Counts" as abundance unit is dimensionless.

Re: Peak hight

Posted: Tue Apr 10, 2012 12:10 pm
by Peter Apps
Hi Jaap

"counts" is how many ions hit the detector. The only applicable SI unit is the Mole.

Peter

Re: Peak hight

Posted: Wed Apr 11, 2012 2:47 am
by wulisheng2
Hi Jaap

"counts" is how many ions hit the detector. The only applicable SI unit is the Mole.

Peter

Hi,
I guess the "counts" is a part of term "Total Ion Counts (TIC)". If yes, there is no unit for "counts" then. It is just a natual number. However, the y axis units, ABUNDANCE, for either TIC, (time vs. abundance) or mas spectra (m/z vs. abundance) uses unit of percentage or ppm (1% = 10,000ppm).