5890 Flow Rates - what am I missing?
Posted: Thu Mar 01, 2012 5:12 am
Hi,
I've worked as an operator on HP 5890s and 6890s but this is the first time I've had to set one up from scratch and I'm missing something in setting the flow rates. I'm working with 5890 II with EPC and an NPD. The column is a Qt Bond PLOT, 30m x 0.32 um. I'm trying to match the following parameters, using Helium as the carrier and makeup gas:
FLOW RATES
Carrier (He) 20 +/• 1 ml/min
Hydrogen 3.5 +/• 0.1 ml/min
Air 100•120 +/• 10 ml/min
Makeup (He) 10+/•1 ml/min
Split Vent N/A ml/min
Septum Purge 1•2 ml/min
(This is the NPD test method in the back of the 5890 II reference manual.)
Oven isothermal at 180C
I don't understand how to get the the carrier gas to that rate. I can set up the column head pressure to the recommended starting values in the reference manual for my column size (anywhere from 7-13 PSI), but when all other gas flows are off, I only get 1-2 mL/min out of the detector. How am I supposed to hit 20 ml/min other than crank up the PSI to an ungodly high number? When I use the HP flowcalc software it says I should only be getting 1.33 ml/min out the detector.
Also, am I supposed to get nothing out of the split flow for this?
Thanks!
I've worked as an operator on HP 5890s and 6890s but this is the first time I've had to set one up from scratch and I'm missing something in setting the flow rates. I'm working with 5890 II with EPC and an NPD. The column is a Qt Bond PLOT, 30m x 0.32 um. I'm trying to match the following parameters, using Helium as the carrier and makeup gas:
FLOW RATES
Carrier (He) 20 +/• 1 ml/min
Hydrogen 3.5 +/• 0.1 ml/min
Air 100•120 +/• 10 ml/min
Makeup (He) 10+/•1 ml/min
Split Vent N/A ml/min
Septum Purge 1•2 ml/min
(This is the NPD test method in the back of the 5890 II reference manual.)
Oven isothermal at 180C
I don't understand how to get the the carrier gas to that rate. I can set up the column head pressure to the recommended starting values in the reference manual for my column size (anywhere from 7-13 PSI), but when all other gas flows are off, I only get 1-2 mL/min out of the detector. How am I supposed to hit 20 ml/min other than crank up the PSI to an ungodly high number? When I use the HP flowcalc software it says I should only be getting 1.33 ml/min out the detector.
Also, am I supposed to get nothing out of the split flow for this?
Thanks!