GC does not go in standby mode anymore, since it starts up below room temp, LN2 consumption is through the roof.

GC 7890 DHA6729 application by AC
Chemstation B04.02 SP1

Post sequence box checked, Macro "standby.mac", go.

We cannot find standby.mac anywhere on the C drive, is it hidden or is it the reason why GC does not go in standy mode post sequence?...

Where can I get this standby.mac file and must I adapt it to fit my application?

Thanks
Have you tried running a dummy method as the last line in your sequence? One that is always above ambient or with no cryo so your LN2 isn't running.
The macro file should be in the CORE directory of the ChemStation, i.e. C:\Chem32\Core. If it is not there, I would think that AC probably supplied a CD with the macro on it. If you do not have this macro, you should contact AC.

Gasman
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry