Page 1 of 1

double side gold seal

Posted: Wed Feb 29, 2012 2:39 am
by edmgc7788
hi, my company tried to save money so now i am testing the double side gold seal. But the tailing is more bigger than before. Is the ferrul the same or should be bigger? My instrument is 6890NB, GCFID, use the temperature programme
Is there any suggestion for me how i can solve the problem?
thank you
rebecca

Re: double side gold seal

Posted: Wed Feb 29, 2012 2:47 pm
by Consumer Products Guy
You mean the double-vespel inlet gold seal, like from Restek?

We use them, and we like them.

My guess is an installation issue. Re-do the installation, and that seal is re-usable. How far does the capillary protrude fast the ferrule? Did you trim the column a bit after inserting through the ferrule? Stuff like that.

Re: double side gold seal

Posted: Wed Feb 29, 2012 5:45 pm
by aldehyde
If the only difference is a bottom vespel seal rather than a washer I don't think the gold seal is the issue. Your column positioning or some other inlet problem is more likely.

Re: double side gold seal

Posted: Thu Mar 01, 2012 1:57 am
by edmgc7788
Yes, I did trim the column. The column is new, before i tried the double side gold seal, it's good. I wil try to reinstall again, maybe the ferrule is wrong.
when you flip to other side, the surface is ok? mine is dark.
thank you guys to answer my question