Baseline plateau GC/HS injection
Posted: Tue Feb 28, 2012 7:39 pm
During GC/Headspace runs lately I have been seeing a "plateau" where the baseline rises at around 2.4 min, flattens out and then falls back to where it originated at about 3.4 min. This is occuring during my water diluent injections and when any other solution water based is injected. I am not seeing it if I inject a blank vial through the headspace or on a direct injection not using the headspace on the same columns. I have eliminated the column as the source by trying 3 or 4 different ones. It looks to be coming from the headspace. I have changed out the sample loop and sample probe, and nothing has changed. I am running the USP 467 Residual Solvent method on a G43 column (30m x 0.53 mm x 3.0 um). The GC conditions are as follows:
Chromatographic Conditions
HP 6890 GC with G1888 Headspace
G43 capillary
Flow: 4.0 mL/min
Inlet: 140°C Split Ratio: 5:1
Detector: 250°C
Oven: 40°C, 20min to 240°C@10°C/min, 20min
Headspace Parameters(USP<467> set 1)
Oven: 80°C Loop Fill: 0.5 min Vial Size: 20 mL
Loop: 85°C GC Cycle Time: 70 min Inj/vial: 1
Tr. Line: 85°C Loop Eq Time: 0.1 min Shake: Low
Vial Eq: 60 min
Inject: 1.0 min
Press. Time: 1.0 min
Any suggestions would be appreciated.
Chromatographic Conditions
HP 6890 GC with G1888 Headspace
G43 capillary
Flow: 4.0 mL/min
Inlet: 140°C Split Ratio: 5:1
Detector: 250°C
Oven: 40°C, 20min to 240°C@10°C/min, 20min
Headspace Parameters(USP<467> set 1)
Oven: 80°C Loop Fill: 0.5 min Vial Size: 20 mL
Loop: 85°C GC Cycle Time: 70 min Inj/vial: 1
Tr. Line: 85°C Loop Eq Time: 0.1 min Shake: Low
Vial Eq: 60 min
Inject: 1.0 min
Press. Time: 1.0 min
Any suggestions would be appreciated.