Hi, I’m developing my first HPLC Method and I can’t seem to find any information on what is generally accepted as a minimum k’ for any given peak of interest. Any thoughts?
Posted: Mon May 09, 2005 10:21 pm
by james little
LCGC is a great source of information about LC method development. I especially enjoy Dolan's articles. I am a mass spec person, but do a lot of HPLC method development. The information in his articles has been invaluable.
see page 1 of
http://www.lcgcmag.com/lcgc/article/art ... ?id=131256
to answer your question.
When developing LCMS methods, I always make sure I retain the peak several column volumes. Especially in protein precipitated samples. There is often material at the first of the column from the sample matrix that can suppress the signal. I always prefer gradient methods over isocratic ones.
Good luck.
Posted: Mon May 09, 2005 10:32 pm
by cafran
Thanks,
That's exactly what I was looking for.
CF
Posted: Tue May 10, 2005 3:01 pm
by Provetech
I just would like to make you notice that the "new" name for K' is K called "retention factor" formerly named "capacity factor"....

Posted: Tue May 10, 2005 4:58 pm
by Victor
.....er, I thought the retention factor was given the symbol
k
That's small case italic. K means something else.
Posted: Tue May 10, 2005 5:42 pm
by DR
As my chem. teacher used to say, "A 'k' is a 'k' is a 'k' . . . "
Posted: Wed May 11, 2005 9:29 am
by Provetech
....yes, you're right

it's "k "and not "K", but the comment is still available on change of k' to k....
Posted: Wed May 11, 2005 10:12 am
by Victor
Provetech
I'm glad we agree .
k is the IUPAC symbol for retention factor, the artist (!!) formerly known as k'.
However, k or k' should not be confused with K which refers to the partition ratio. The partition ratio in chromatography is related to the retention factor via the phase ratio of the chromatographic system.
So a k is a k is a k' but never a K
Chemistry is confusing with symbols. k is often used for the Boltzmann constant and K for the absolute temperature etc etc. Perhaps we should have adopted Chinese characters instead. There are many more than 26 available.
Posted: Thu May 12, 2005 4:21 am
by tom jupille
Victor, I believe that the USP and ICH documents (which chromatographers in the pharmaceutical industry must follow) still use k'. (This causes some confusion for us in our short courses!)
Posted: Sun May 15, 2005 9:52 pm
by BigGeoff
To reiterate James Little's comments on LCGC and Dolan's articles - they are goldmines of information. My personal favorite is the series 'Starting out right', in , I think 5 parts starting January 2000.