I feel that some component in plasma is interacting/binding with the DHBA, causing it to read higher than it should be (about 5-10X higher than what it should be).
Any thoughts
Here are my extraction methods using solid phase extraction with a vacuum manifold:
1) Condition Supelclean LC-WCX tube with 500 ul 0.5 hydrochloric acid.
2) Wash tube with 1 ml DI water to remove excess acid.
3) Load sample (500 ul plasma sample with 500 ul DI water), pass through tube at 250 ul/min.
4) Two washes (1 ml each) using DI water.
5) Add internal standard (50 pg DHBA in 250 ul water), pass through tube at 250 ul/min.
6) Elute with 250 ul perchloric acid.
7) Injection into Eicom H-TEC 500 is 20 ul of eluent.
Again, any ideas on why some plasma samples have large spikes in DHBA and others do not?