Missing Methane
Posted: Tue Feb 28, 2012 2:36 am
by elleraim
Hello!
I am a novice graduate student hoping to find some help. My goal is to use a Varian Saturn 2100T GC/MS to find methane in our gas standards; a mix of air, methane and CO2, used as a set up for future methane analysis. So far, no peaks of methane have shown up but O2 and CO2 create peaks. The column is capillary and it was recommended I use a PLOT (25m) column. Do I need to purchase a new column? Or can the GC/MS detect the methane as is? Any advice would be great. Thanks.
Re: Missing Methane
Posted: Tue Feb 28, 2012 3:26 pm
by AICMM
elleraim,
The forum has lots of suggestions about not using MS for gas analysis. You should take a peek at these to get a handle on some of the issues. Very hard to do well for lots of reasons. (My standard disclaimer applies here but the advice still holds.) Is there any way you can possibly do this by GC? Do you have an available FID electrometer for your GC?
Best regards,
AICMM
Re: Missing Methane
Posted: Tue Feb 28, 2012 4:53 pm
by elleraim
Unfortunately the only machine available for use is the GC/MS. I have noticed that a GC-FID is the most common, but I was hoping to work with what we have. Thanks so much for the advice. I will comb the topics more.
Re: Missing Methane
Posted: Tue Feb 28, 2012 10:16 pm
by npaz
Make sure you have the right column. Perhaps a nonpolar RTX-5 from Restek. Second with what you are describing I would recommend SPME ( solid phase microextraction ). Set the fibers in your sample with a tight cap with septa tops. They are fairly easy to use. When you run this with a GC/MS do not use a solvent delay set it to 0 minutes since your gases will elute well below a minute.
Re: Missing Methane
Posted: Wed Feb 29, 2012 7:28 pm
by Bigbear
Make sure you're scanning "low" enough. In my world ( Agilent) the default low mass is 50. You may want to set uo a sim method or scan a naarrow range.
Re: Missing Methane
Posted: Wed Feb 29, 2012 10:54 pm
by chromatographer1
You can check with Supelco about SPME for methane, but if I remember correctly, it is not feasible.
best wishes,
Re: Missing Methane
Posted: Thu Mar 01, 2012 12:26 am
by nvalentine
Check with the column manufacturers, but from what I recall, they do not recommend using PLOT columns on a mass spec.
A normal phase column (i.e. DB-5, RTX-5) will not separate that light of gases. You definitely need the phase from a PLOT or some type of molecular sieve to do it.
An FID or TCD would be ideal.
Re: Missing Methane
Posted: Thu Mar 01, 2012 1:46 pm
by Bigbear
Oh contrare, we use a 30M X 0.32 plotQ column for a CDC CN method using a 5973.
Re: Missing Methane
Posted: Thu Mar 01, 2012 2:20 pm
by AICMM
elleraim,
If you are stuck with MS, my best suggestion is a really thick film DB-1/SPB-1/RTX-1 type column. I would also check to see if your GC has the option of cryo (CO2 or LN2) or look at the possibility of throwing a piece of dry ice in there and see if it hits a low temp set point.
Bigbear, what is CDC CN?
Best regards,
AICMM
Re: Missing Methane
Posted: Thu Mar 01, 2012 3:41 pm
by nvalentine
Yes, I had a conversation previously with Restek and they said you should be safe with a diameter less than 0.53 on a PLOT for mass spec. Safe, but I got the feeling not "recommended".
Re: Missing Methane
Posted: Thu Mar 01, 2012 6:42 pm
by Bigbear
CDC Center For Disease Control CN cyanide
Re: Missing Methane
Posted: Fri Mar 02, 2012 3:25 pm
by AICMM
So there's two issues with PLOT's and MS. The first is that they sometimes spit off column. Much better than the bad old days but still happens once in a while. This lands somewhere and that somewhere is your source. May or may not be an issue. The second issue is that a 0.32 optimal flow is somewhere between about 2 and 5 mL/min but most MS's only like about 1 mL/min. If your MS can handle more (the newer MS's with turbo's often can), then a 0.32 might work okay, if only 1 mL, then the plot is not operating at optimum and the curve is very steep on that side. But what can you do, right?
Bigbear, regarding CN, you mean HCN I assume?
Best regards,
AICMM
Re: Missing Methane
Posted: Mon Mar 05, 2012 2:00 am
by nvalentine