questions for seal pack/needle replacement

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi,

I recently found that the seal pack from my 2695 separation module is leaking. The seal pack rebuilt kit has not arrived yet.

I was reading the 2695 manual and got a few questions.

1. Is it okay to rebuild the seal pack first, and replace the needle after that? Maybe I got confused between needle replacement and reinstalling needle, but from the Waters video, seems that reinstalling the needle looks the same as needle replacement in the first part (aligning the needle, placing the needle into the needle tee etcs). I thought that it would be okay to rebuild the seal pack first, and replace the needle right after... Could anyone correct me if I am wrong?

2.To perform the compression check. Should I use a pin plug or column cap to tighten the column outlet to generate a back pressure upto 500psi, or just let the column outlet open and collect the solvent into a reservior?

Questions look stupid but I would really appreciate if anyone can answer me patiently. Thanks a lot!

Hyuna
Yes, installing the needle would be the same procedure whether it's the same needle or a new one. A lot of the time and effort of rebuilding the seal pack is taking the assembly out and putting it back, so re-installing the same needle and then replacing it shortly thereafter would be a lot of extra hassle.
David Hobbs
Instrument Services Specialist
Pace Analytical Services, Inc.
david dot hobbs at pacelabs dot com
Hi,

1. Perfectly okay to rebuild the seal pack and then insert the needle. Need the ensure the new needle is seated correctly or you'll create a void.

2. A column or restrictor is okay to use. Ensure you pressurise the system up too at least 4000psi to bed in the seals for ten minutes.

Regards,
Bull76
Hey Guys,

Thanks a lot for the advises and the new seal pack is working perfectly okay after the replacement 2 weeks ago. I didn't follow the instruction as I thought it would cause too much hassle. I was glad that you guys had the same thought as me and I didn't really mess up my second maintenance.

Cheers,

Hyuna
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry