Maintenance of RF 18 column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I am attempting HPLC for the first time.
I have used an RP 18 (Agilant Zorbax ODS C18) for separation of ascorbic acid with mobile phase sodium acetate , EDTA , acetic acid . However I could not wash the column post use. It lying like this disconnected from the system for a month now.Is there any chance of contamination or other problems . I would use it within next 7 days. Do any washing or any treatment I need to do befor next use.
Kindly suggest
A column that has been allowed to dry out can sometimes be recovered, but there is no guarantee. A reverse-flush with a 50:50 mixture of methanol and water at a fairly low flow rate (maybe 0.5 mL/min if it's a 4.6-mm id column) is often a good starting point.

However, you should check with Agilent to see if they have any specific recommendations. If they do, follow them.

In any case, be prepared to buy a new column.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 40 users online :: 2 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry