pressure problem
Posted: Wed Feb 22, 2012 9:26 am
by Pietn23
I'm using an agilent 6890 GC-MS.
When I want to turn up the inlet pressure to about 80 psi it goes up very slowly until 50 and then stops.
I already changed the column, replaced the septum, liner and ferrules.
Anyone has had this problem before?
Re: pressure problem
Posted: Wed Feb 22, 2012 10:06 am
by Peter Apps
Yes, this turns up frequently on the forum (and is sitting in the archives ready for someone to search for it), usually the problem is a leak or a blocked split vent line and filter.
Posted: Wed Feb 22, 2012 1:17 pm
by aldehyde
Could also be insufficient line pressure, if your carrier gas regulator is set to only supply 50 psi. Also if it is increasing slowly you probably either have gas saver on or a low split. Set it to split mode with a split flow of like 200 mL/min and it will increase much more quickly.
Posted: Wed Feb 22, 2012 7:05 pm
by uzman
80 psi is a very high pressure for GC ( more than 5 bars ) , are you using a very long column ( over 100 meters ) ?
I think this pressure is out of control limit for EPC.
Posted: Fri Feb 24, 2012 8:45 am
by Hornet
Can i ask you why you need a pressure of 80 psi? :shock: