Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
We have a dirty flow cell on our Cap LC and cannot get it clean. We have tried 100% H2O and 100% MeOH (overnight) but the back pressure is still very high.

Can anyone give some suggestions of what to try next?.

For info - we have been running biological samples in ammonium formate/acetate + MeCN / MeOH mobile phase.

Thanks all!.

James

Try forcing dilute nitric acid in the flow cell by manual glass syringe.

JM

If there is a possibility of having proteins, etc., in there you may want to use detergents, maybe with dithiothreitol, or urea (or other chaotiopes) first.
Thanks, I'll pass that on to our Cap-LC analyst

James
tayzyboy,

Did you try reversing the flow to the flow cell? This doesn't always work but often a partial blockage is effectively removed by simply reversing the low for a few minutes and then restoring flow to the normal flow direction.
5 posts Page 1 of 1

Who is online

In total there are 24 users online :: 3 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry